Question 1: Which concentrations of ozone are required in water systems?
The technical literature delivers different information about the ozone concentrations in water systems: e.g. ISPE Baseline Water and Steam: 0.02 ppm - 0.2 ppm; Collentro, Pharmaceutical Water: 0.2 ppm - 0.5 ppm and W.Setz, Ciba-Geigy 1990: max 0.04 ppm, for sanitisation 0.05 ppm.
The indications provided by the ISPE Baseline refer to the concentration required to prevent microbial growth. One can thus assume that a concentration of 20 ppb ozone can prevent any growth.
If systemic protection is desired i.e. the constant presence of ozone in the water, lower ozone values are sufficient.
In practice, approx. 0.02 to 0.05 ppm should be sufficient for Aqua Purificata. For sanitisation, it naturally depends on the sanitation time intervals - daily or weekly. Finally, the required ozone concentration for the system should be determined within the framework of the validation for the whole system.
http://www.gmp-compliance.org/en ... tical-Water%22.html |
|