最新的1058可是说了B类可能只需要校准,可能需要IQ、OQ,可能需要全验证。要依据实际情况评估。扯了这么多直接上原文:Instruments can generally be classified as belonging to Groups A, B, or C. It should be noted that the same type of instrument
can fit into one or more categories, depending on its intended use.
Group A includes the least complex, standard instruments that are used without measurement capability or user requirement
for calibration, such as a magnetic stirrer or vortex mixer. Proper function is ensured by observation, and no further qualification
activities are needed for this group.
Group B includes instruments that may provide a measurement or an experimental condition that can affect a measurement.
Examples include a pH meter or an oven. Proper function of instruments in this group may require only routine calibration,
maintenance, or performance checks. The extent of activities may depend on the criticality of the application. Generally,
these instruments may have firmware but not software that is updated by the user.
Group C comprises analytical instruments with a significant degree of computerization and complexity, such as high-pressure
liquid chromatographs and mass spectrometers. All elements of qualification, including software validation, must be considered
to ensure proper functioning of instruments in this group.
这是最新的1058,你帖子里面的是以前的。
|
|